$\begingroup$

I am reading this journal article which discusses using high-performance liquid chromatography (HPLC) to separate dyes from wine. Here are the relevant facts about the experimental setup:

  1. Two mobile phases are used, a solvent A and a solvent B. Solvent A is is 40% methanol and solvent B is 70% methanol. Both solvents also have 25% 0.02M tetrabutyl-ammoniumhydroxide phosphate solution as well.
  2. The stationary phase is a LiChrospher 100 RP-18e column.

Below are the chromatograms and the retention time tables that this experiment produced: Chromatogram of various wine colorants

Table of elution times of various wine colorants

From the chromatogram, we see that amaranth has the fastest elution time of 6.60 minutes. My questions relate to what inferences can we draw from elution times:

  • Does the elution time correlate with the polarity of the compound itself? Here, it states that similarities in polarity correlates positively with elution time. The stationary phase used in this experiment (the LiChrospher) is a nonpolar silica column. Thus, with amaranth coming out the fastest, would it make it the most polar compound?
  • Further, how does the use of two solvents (A and B) affect making an inference like the one above? In the chromatogram above (specifically chromatogram "a") and the associated table, the elution time for the three compounds ranged from 6.60 minutes to 9.76 minutes. Since a solvent A and a solvent B were used, how do I know which compounds were eluted by which solvent? (If my interpretation of how the solvents A and B work is wrong, please correct me). Also, how does the use of two solvents affect understanding the relative polarity of the compounds?
$\endgroup$

1 Answer 1

2
$\begingroup$

You have several queries. First of let us look at the gradient method used by the authors.

The HPLC gradient for synthetic colorants was 0 min A 100 p. cent, 10 min A 0 p. cent, 14 min A 100 p. cent. The HPLC gradient for natural carmine was 0 min A 80 p. cent, 20 min A 50 p. cent, 23 min A 30 p. cent, 25 min A 30 p. cent

Factually, notice the trend in the mobile phase for synthetic ones. It is a linear gradient elution that they started with 40% MeOH at t=0 min, and at t=14 min, you have 70% MeOH. In the second case, it is a purely empirical "step" gradient. Try to draw a graph on a piece of paper.

Regardless, the authors use a water-rich mobile phase in the beginning and go to methanol rich mobile phase. Note I am avoiding the word polarity.

  1. Polarity is an umbrella term in chromatography. For elementary purposes, it may be okay for serious work predicting retention on the ill-defined polarity that carries little or no meaning. You linked a GC webpage for an HPLC separation; those are different worlds. Qualitatively, in top figure (a), it is clear that amaranth spends more time in the water-rich mobile phase, followed by azo ruby and then ponceau 4R if we use a C18 phase.

All you can say is that amaranth has a higher solubility in the water-rich mobile phase than ponceau 4R. One should not over-interpret retention behavior with this limited data.

In short, polarity is not a great terminology for describing the retention order. If you are interested, see the "Hydrophobic subtraction model" of Snyder et al. This one describes the retention in a better way of C18 molecules.

2)

In the chromatogram above (specifically chromatogram "a") and the associated table, the elution time for the three compounds ranged from 6.60 minutes to 9.76 minutes. Since solvent A and solvent B were used, how do I know which compounds were eluted by which solvent?

This is an interesting question. As you noticed, the composition of the mobile phase changing with time, so the k, the retention factor is changing with time. People have worked out the theory of simple linear gradients. You know the retention time of elution, and you have the gradient composition chart. Correlate it, but this is a very, very rough estimate. The actual treatment is way more involved. Check "Theory of Gradient Elution Liquid Chromatography with Linear Solvent Strength: Part 1. Migration and Elution Parameters of a Solute Band" by Leonid Blumberg to see the complexity of the problem of gradient elution.

One of the best books is "Introduction to Modern Liquid Chromatography By Lloyd R. Snyder, Joseph J. Kirkland, John W. Dolan, 2011."

$\endgroup$